Hello i hope everyone is having a great time at SIP now. its been so long now. For today i'll be blogging on osmolality. and my lab uses a different equipment from our school's. it seems to be more advanced. well on to it =)
Osmolality is an expression of total concentration of dissolved particles in a solute without regard to the particle size, density or electrical charge. The measure of total solute concentration can only be made indirectly by comparing one of the solution colligative properties with corresponding cardinal property of pure solvent.
Vapro osmometer is based upon a measurement of vapor pressure depression made possible by thermocouple hygrometry. This sensitive temperature sensor operates on the basis of a thermal energy balancing principle to measure the dew point temperature within the chamber.
10 micro liter of specimen is aspirated into a micropipette tip and then inoculated into a solute free paper disc. The paper disc is then pushed into the instrument and sample chamber is locked. The sensing element is a fine wire thermocouple hygrometer. This is suspended in a unique, all metal mount, which forms a small chamber enclosing the specimen when joined with the sample holderAs vapor pressure equilibrates in the chamber airspace, the thermocouple then seeks the dew point temperature within the enclosed space, giving an output proportional to the differential temperature. The difference between ambient temperature and dew point temperature is the dew point depressionSample used- serum or urineStepwise.push sample chamber level upwards and pull the sample slide out from the instrument.
use forceps supplied to place a single sample disc in the central depression of the sample holder. If more than two discs are used, results will slightly elevated
Using a micropipette, pipette 10ul of sample
Rest pipette tip on the notch of pipettor guide. It should be about 5mm away from center of sample discsmoothly depress micropipette plunger to release sample onto sample discwith plunger still held down, lightly touch the pipette tip to sample disc, then lift it away. Tip must briefly contact sample disc to press the disc flat against the holder
Sample disc should appear fully saturated, with a slight liquid meniscus on surface.
Note: when cleaning the sample chamber, use only lint free tissue. Facial or soft tissue will produce excessive lint residue that will contaminate the thermocouple sensor
Yuxuan
Sunday, August 3, 2008
Subscribe to:
Post Comments (Atom)
7 comments:
hey yuxuan
if it is possible, can you explain the terms colligative property and cardinal property coz don't really understand.
thanks for your time
rusydiana
hey..
do you do QC for osmolality?
if you did, how many level of QC is dat?
sutiana
Hi,
May i know what kind of solutes do you deal with most of the time? Also, can you explain more about thermocouple hygrometry? Is it an equipment or a theory?
Thanks
Ka Hang
TG02
Hi Yuxuan,
1)Do you mind explaining this sentence: "The measure of total solute concentration can only be made indirectly by comparing one of the solution colligative properties with corresponding cardinal property of pure solvent."
2)"Vapro osmometer is based upon a measurement of vapor pressure depression made possible by thermocouple hygrometry."
-What is vapor pressure depression and thermocouple hygrometry?
3)"As vapor pressure equilibrates in the chamber airspace, the thermocouple then seeks the dew point temperature within the enclosed space, giving an output proportional to the differential temperature. The difference between ambient temperature and dew point temperature is the dew point depression"
-Do you mind explaining and simplifying the above?
4)What are the results obtained? Any reference range to it?
5)When is osmolality testing necessary to be performed?
Thankz!
Han Yang
TG01
Please reply to your friends questions ASAP. You need to reply within 2 weeks from your date of posting.
to sutiana: yes we do run and we do run controls too.
to ka hang: we recieve CSF serum and urine samples. and that is a theory which the machine runs on
han yang: vapor pressure depression is the method of the new osmometer in the lab like the one in our sch is called 'freeze point depression'
and for the huge chunk, it basically means in the chamber to measure the solute, dew point temperature is measured, giving a differential temperature same to the dew point temp.
results obtained can vary. urine for example have very high ranges